Multiplexing PT100s
14 posts / 0 new
I wanted to interface 8 PT100s (2-wire) to ATMega64A. How can I compensate for the "ON" resistance introduced by the MUX? By using a MUX I can just use 3 comparators which will make my PCB smaller.
If you have to stick with a 2-wire connection then it will be difficult.
What mux were you planning on using?
"How have you proved that your chip is running at xxMHz?" - Me
"If you think you need floating point to solve the problem then you don't understand the problem. If you really do need floating point then you have a problem you do not understand." - Heater's ex-boss
Depends on how you're reading the Pt100s. If you use a constant current source then read the voltage across the Pt100, then the trick is to ensure the mux is not part of the current path - it only switches voltage.
To the OP - PT100 temperature sensors have a quite small (but linear) temperature coefficient. It is not easy to measure with a low resolution ADC. Note that the ADCs in that diagram, above, are 24 bit! When you add multiplexing, it becomes a really big challenge. Better, 8 interface ICs that you read serially (TWI or SPI).
will it really make it that much smaller...you invite many headaches....if so, take Kartman's advice, mux only the reading of the voltages
Thank you for the reply everyone.
I was thinking of using CD4067 or similar, had a look at ADG708 as it has low "ON" resistance of about  3 ohms but is very expensive while CD4067 has an "ON" resistance of about 56 ohms at 5V.
Yes I am switching the voltage, this is what I am trying to do.
Mulitplexed PT100s
Yes if it is getting complicated I might opt out of it but eventually I will have to try this as the number of sensors will keep increasing.
What is the 4052 feeding? More importantly, what impedance is it feeding?
So why are you concerned about the on resistance of the mux? The source is low impedance and the adc input is high impedance. 60 ohms vs 600MOhm is not much is it? The humble 4051/2/3 or 4067 is most likely more than adequate for your task.
The schematic says Pt1000. And why a 100uF cap?
PT100 or PT1000 has very few ohms delta for temperature change and the impedance for the digital mux will give you a big error for measure.
use a bridge methode measure with oamp and after mux the output of oamp.
Thierry
Thierry Pottier
There should be more wires going to the sensor through the mux...your current drive, & the sense lines (+ and -).
Your arrangement will give pretty horrible accuracy.  If you are only trying for low accuracy, just use an LM35, or thermistor.
http://www.ti.com/lit/ug/slau520a/slau520a.pdf
http://www.ti.com/europe/downloads/2-%203-%204-Wire%20RTD%20Measurement.pdf
or other cheap sensor
Digital sensors often have bus communication - so no need for any (hardware) multiplexing!
Kartman wrote:
So why are you concerned about the on resistance of the mux? The source is low impedance and the adc input is high impedance. 60 ohms vs 600MOhm is not much is it? The humble 4051/2/3 or 4067 is most likely more than adequate for your task. The schematic says Pt1000. And why a 100uF cap?
Yeah I will go ahead with it. The diagram i posted is from the article I am using to interface the PT100 sensor.
https://learn.openenergymonitor.org/electricity-monitoring/temperature/rtd-temperature-sensing?redirected=true
AVR
Yes, I was afraid of exactly this, hence I started this post.
I am using LM35 currently. The problem is that they are not easily available in the market whereas PT100 is available everywhere, so I decided to switch.
Thank you for the links, they are brilliant, specially for a beginner like me.
I will post on this when I get my results.
Thank you all.
What kind of system are you building?
How far are the sensor apart?
You might want to consider to make "smart" sensors. Just use an AVR for each sensor and do (long) wires with RS485 (modbus) or similar.
Paul van der Hoeven.
http://www.hoevendesign.com